Reviewing my posts I may not have been clear about the additional samples run at the same time as the std addition samples.
The dissolution solvent for the additional samples must be the SAME as for the std addition samples.
So, as in the example I provided, the drug samples must also contain 100µL of solvent AND 10µL of DMAc containing 0 ppm of the residual solvents that were added to the other samples. The weight of the drug must also be the same (5mg).
The pure dissolution solvent then consists of 110µL of the 10:1 co-dissolution solvents, water and DMAc.
I hope I caused no confusion.
It is also possible that one of the std additions can be 0 ppm. Thus only 2 std addition solutions are prepared and a zero ppm addition (pure DMAc) is one of the three points, but I would prefer that three std addition solutions are prepared. This makes the procedure more rugged and more likely to prove errors clearly if any are made (non-homogeneity of the samples or errors in the preparation of std addition solutions, which of course, never happen in the real world, right?
best wishes,
Rod