injector temperature

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
Should my injector be hotter than the maximum column temperature?

Im using a J&W DB 624 column with a maximum usable recommended column temp of 260C, weve been hammering it at 275C for over month now!

Im sure its affecting my results!

It is not generally a good idea to heat a column above the maximum temperature for extended periods of time. 275 is pretty hot for that column, but it is a relatively thick phase and will take longer to burn off the phase and get down to bare column, which is a very bad thing in many cases. I would reduce the injector temperature. If you have to exceed the recommended maximum temperature limit to vaporize the compound it probably will not travel through the column well and deposit heavy residues which will affect column performance. I would clip off a foot of the column and lower the injector temperature.

I agree with Ron. However, if for some reason you need to have your injection port above the column's max temp, you should consider using a short piece of guard column, which typically has a max temp of ~325C.
Answer Man
3 posts Page 1 of 1

Who is online

In total there are 196 users online :: 1 registered, 0 hidden and 195 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Semrush [Bot] and 195 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry