Helium to Hydrogen Method Transfer Problem
Posted: Mon Sep 26, 2011 2:51 pm
Hi folks,
I have a method for a FAME (fatty acid methyl ester) mix that is well established in our lab using helium carrier. Recently, we bought a couple new Agilent 7890 GC's and decided to set one up with hydrogen carrier for now, with the idea in mind that we will eventually be moving in that direction. The FAME method is one we run a lot and that I would like to get transferred over to hydrogen first. It's ideal because we still have to perform the validation, so we can demonstrate equivalency now, at a convenient time. We have no experience in working with hydrogen carrier, however.
I had some samples prepared and divided between the two 7890's, including a previously-analyzed lot, so that I could compare exactly the same solutions with helium and hydrogen carrier on the two new, otherwise-identical systems.
On the 7890 running helium carrier, the previously-analyzed sample gives exactly the same results as on our older 6890 systems.
For the system running hydrogen carrier, I tinkered a lot with it, and found that I could almost exactly reproduce the selectivity and resolution, but with half the run time. What seemed to work was just scaling the oven program (doubling the ramp rate and halving the hold time), and then playing with the initial linear velocity until a key reference peak eluted at the "expected" analog of its place in the helium chromatogram, at the "top" of the oven ramp.
The only problem is that in the hydrogran chromatogram, there is a lot of small "junk" close to the top of the oven ramp over a space of several minutes that is not present in the same sample run on the 7890 with helium carrier, or the original release chromatogram. The blank diluent injections do not contain this "junk," and the column compensation runs are extremely clean, but all of the FAME samples, regardless of origin, contain the same junk.
This is a deal killer for me, because my method requires me to identify a large number of very tiny peaks, and quantify by area percent. I'm thinking of changing the carrier to helium on this system and giving up for now because I'm very pressed for time, but I was hoping that if I put my problem out there, someone with more experience in working with hydrogen carrier might be able to tell me something to look for.
Thanks for any advice you can offer,
Stephen
Stephen