Page 1 of 2

sec baseline

Posted: Thu Sep 22, 2011 1:55 pm
by MestizoJoe
I'm having serious baseline issues with SEC. The column is new and the mobile phase is 30% methanol 70% phosphate buffer. There is no noise without the column. The phosphate buffer is filtered through a 20 micron membrane. The column guard doesn't make the noise better or worse.

Image

Lately the noise has been worse than this.

Thanks.

Re: sec baseline

Posted: Fri Sep 23, 2011 4:04 am
by unmgvar
can you say the type of the SEC?
MeOH at these amounts is not recommended for use in a lot of SEC columns

Re: sec baseline

Posted: Fri Sep 23, 2011 7:26 am
by Gerhard Kratz
Standard shipping solvent is 0,05% NaN3 + 0,1M Na2SO4 in 0,1M phosphate buffer pH 6,7! What is your buffer concentration? Is it a silica based SEC column? Salt concentration should be <0,5 M! When changing solvent I would recommend a flow rate at 0,3ml/min. What is your flow rate and what back pressure do you monitor.

Re: sec baseline

Posted: Fri Sep 23, 2011 7:16 pm
by nschwartz
It is important to know what type of SEC column this is and whether the chromaogram you are showing is with or without an injection. Also, how much is this noise relative to the response you get for your analytes?

Re: sec baseline

Posted: Fri Sep 23, 2011 7:34 pm
by MestizoJoe
Hi everybody. The SEC columns used are biosep-sec-s4000 followed by biosep-sec-s2000 with pH 7 and 50 mM phosphate. Its 70% buffer 30% methanol.

the UV looks fine but the RI has very high noise. peaks from noise are like 10% the area of the sample peak (approximately). there is no noise with just water.

this is at 248 nm also. solvents degassed by He sparge and there is a degasser on the instrument.

we've tried:

filtering buffer through finer membrane (0.2 micron)
getting new columns (columns now are brand new)
long equilibration times
using a precolumn/column guard (made no difference in baseline)
yelling at problem (good stress relief; no closer to solving problem)


Could decreasing the buffer concentration help?

SEC retains small molecules (like buffer salts). could this be a result of buffer salt retention and elution not being completely uniform?

Re: sec baseline

Posted: Sat Sep 24, 2011 5:47 pm
by unmgvar
are you premixing or using 2 bottles?

Re: sec baseline

Posted: Sat Sep 24, 2011 7:40 pm
by MestizoJoe
premix.

Re: sec baseline

Posted: Sun Sep 25, 2011 9:08 pm
by XL
You said the UV was fine but the RI had high noise. If you run the method without injection, will the RI trace be still noisy? It will be good if you can post the chromatograms with and without injection.

Yes, the smaller the analyte, the long it is retained on a SEC column. Therefore, if your sample contains a host of stuffs such as salts, surfactants, and etc in addition to the analyte, the RI trace will be much more complex than the UV trace.

Re: sec baseline

Posted: Tue Sep 27, 2011 4:00 pm
by MestizoJoe
So when I use waters 2695 with RI detector there is not nearly as much noise as with agilent 1100 series with RI detector.

Re: sec baseline

Posted: Thu Sep 29, 2011 2:58 pm
by unmgvar
is your instrument next to ventilation?

Re: sec baseline

Posted: Thu Sep 29, 2011 3:12 pm
by MestizoJoe
Yes. The ventilation makes no difference however. The waters 2695 with RI detector is right next to it as well and the baseline is fine.

There is some difference between the agilent with RI SEC setup and waters with RI SEC setup. The baseline is fine on the waters system but on agilent system it is horrible. I know waters is like a cadillac and agilent is like a dodge but agilent ought to still produce a stable baseline.

The agilent RI detector is also fine when only water is passing through, but once the 70% pH 7 phosphate buffer 30% meoh solution is passing through there is noise. This doesn't happen on the waters instrument.

Re: sec baseline

Posted: Thu Sep 29, 2011 11:13 pm
by jonh29
Purging the RI is a must. Possibly something to look into. If both systems have been purge adequatley then it's just the sensitivity of the instruments. My experience Agilent instruments provide more sensitivity.

Re: sec baseline

Posted: Fri Sep 30, 2011 4:53 am
by unmgvar
i would check some of the settings on both systems in their methods.
maybe there is a difference that makes the noise on the agilent.
is it only a noise problem, or is the drift also higher?

Re: sec baseline

Posted: Mon Oct 03, 2011 8:52 pm
by MestizoJoe
The dift doesn't seem to be a problem. Both instruments have had the reference cell purged adequately as far as I know.

Re: sec baseline

Posted: Tue Oct 04, 2011 9:45 am
by unmgvar
what about the settings in the instrument method?
are they equivalent?
is the sampling rate the same?