Error 204 for Agilent 5975B inert msd

Posted: Mon Sep 19, 2011 10:36 am
by crazycindy
Hi! We have an Ailent Technologies 5975B inert MSD connected to Agilent 6890 Network GC System. Upon, boot up, a message would appear: "No response, cannot bring MS on line(possible bad address or existing lock [error 204]". I think it has something to do with the communication (probably) of the software and the gc system. Any thoughts on this? Thanks! :) -cindy

Re: Error 204 for Agilent 5975B inert msd

Posted: Mon Sep 19, 2011 9:40 pm
by Steve Reimer
Your computer is not talking to your mass spec. Reboot the mass spec and see if it will communicate. You shouldn't have to vent.
If that doesn't work you may have a board problem. I have a 5973 that has done this twice, the second time it required a smart card replacement.

Re: Error 204 for Agilent 5975B inert msd

Posted: Tue Sep 20, 2011 1:23 am
by willnatalie
Have the bootp program open so that you can watch and see if the MSD IP comes up.

Re: Error 204 for Agilent 5975B inert msd

Posted: Fri Sep 23, 2011 2:44 pm
by Bigbear
You may also want to re boot your ethernet switch.

Re: Error 204 for Agilent 5975B inert msd

Posted: Tue Oct 04, 2011 6:52 am
by crazycindy
Thank you very much for your replies.
We reboot the MS without venting and everything got back to normal again.
Cheers! :)

Re: Error 204 for Agilent 5975B inert msd

Posted: Sat Apr 14, 2012 12:46 am
by laurapast
I have the same problem, but I can not open ChemStation. We check the network cables, communication between the ms and gc with the computer. Try to repair the software, then reinstall the software and still does not open the program. Help please.

Re: Error 204 for Agilent 5975B inert msd

Posted: Mon Sep 10, 2012 7:05 am
by crazycindy
Hi Laura!
What do you mean when you say you can not open the ChemStation? :alien:

Re: Error 204 for Agilent 5975B inert msd

Posted: Wed Nov 21, 2012 2:59 pm
by evagisselle
I have the same problem and also Chemstation can't be opened .... I don't know how to restar the MS without venting... somebody can help me?

Re: Error 204 for Agilent 5975B inert msd

Posted: Wed Nov 21, 2012 7:27 pm
by MSCHemist
You can turn the MS off for a second or two without venting. just hit the power switch. wait a few seconds and turn it back on. What do you think happens in a brown out?

If you have a 5975 and a 6890N there should be no need for BootP. That was for older instruments with jetdirect cards. The 5975 and 6890N are full network capable instruments.

Re: Error 204 for Agilent 5975B inert msd

Posted: Tue Feb 04, 2014 6:55 pm
by howardyu
I just had the same error message on my GC-MS this morning. After re-booting the Ethernet switch, it went normal again.