GC Test Method for Residual Formaldehyde in Herbal Extract
Posted: Sun Sep 18, 2011 1:45 pm
by Terry
Dear all,
I am searching for a GC test method for residual formaldehyde in herbal extracts, any suggestion appreciated.
Terry
Re: GC Test Method for Residual Formaldehyde in Herbal Extra
Posted: Sun Sep 18, 2011 7:02 pm
by chromatographer1
The best method would be to react the formaldehyde with a reagent that allows analysis by HPLC.
Unfortunately after 30 years of trying to analyze formaldehyde in air and in other matrices by GC accurately, I have given up and have decided that LC is a better solution.
Formaldehyde is just too reactive and it even reacts with itself !
I cannot remember the chemical used at this time. You should be able to find it as you seek a HPLC method for formaldehyde. Simply allow the extract to come into contact with the reagent. Then inject the reagent onto a HPLC with UV detector. This reagent is used to measure formaldehyde in air over a period of time.
good luck and I wish my memory was better.
Posted: Sun Sep 18, 2011 8:23 pm
by AICMM
DNPH? There are standards for DNPH derivatized formaldehyde available if I am not mistaken.
I have seen it done directly but it is extremely challenging.
Best regards,
AICMM
Posted: Sun Sep 18, 2011 9:12 pm
by chromatographer1
The LC conditions have to be done just right to separate the derivative from the reagent.
Dinitrophenylhydrazine is the chemical in question.
I believe a tech note was published at Supelco.
best wishes,
Posted: Mon Sep 19, 2011 1:50 am
by Terry
Thank you for your information, I will try to get the reference.
Posted: Mon Sep 19, 2011 5:12 pm
by chromatographer1
Also look at kits to monitor air quality (formaldehyde). You might flow nitrogen/inert gas through your herbal extract to release the HCO into a formaldehyde trap (containing DNPH) which then would react and be extracted and measured by LC.
You could manually spike your extract for recovery studies in this manner.
best wishes,
Posted: Tue Sep 20, 2011 12:40 pm
by JI2002
I agree HPLC is a better approach. There is a report for analysis of formaldehyde in air by GC/MS. The method is not as robust as method TO11.
http://www.entechinst.com/pdf/Formaldeh ... rAir04.pdf
chromatographer1, in addition to reactive nature of formaldehyde, what were the other challenges you had developing GC method?
Posted: Tue Sep 20, 2011 1:42 pm
by chromatographer1
Separation was not an issue, I had no interferences. But the reactivity and non-reproducibility were the real problems. The dimer and trimer of formaldehyde were also present and variable in their content.
Of course, an FID was not a suitable detector. Very low response. And the chromatographic system had to be both clean and inert, both column and sample injection system.
best wishes,