I kindly request your input/comments/best practices regarding a discussion/debate occurring at my workplace. The issue concerns how to spike solvents into an API for accuracy validation.
Assuming the API contains some residual solvents (as they usually do), do you….
1) Spike TO a level, i.e., 100 % of spec. That is, add the appropriate amount of solvent plus what the API contains to equal 100 %.
The rationale for this approach is that if you spike AT the level (described below) and include any residual present the combined response is greater the 100 % (or whatever level you’re testing). Thus, it is not a “true” test of accuracy.
2) Spike AT the level. That is add the appropriate amount of solvent to equal 100 %, then correct (subtract) for any residual present in the API.
The rationale for this approach is, 1) you don't accurately know beforehand the amount of residual present (no validated method) to correctly spike to 100 %, 2) a blank correction is common practice and 3) the linearity validation data would substantiate any combined response
Assumption – the residual solvent is not present in large amounts, i.e., you would not be subtracting a large background from a small signal.