- Posts: 6
- Joined: Thu Sep 08, 2011 3:07 pm
We recently inherited an HP5972 GC/MS system with a 7673 autosampler, running off of a Windows NT 4.0 OS, with ChemStation's "B" series software (G1701BA). The unit has had very little hours on it, but I'm not quite sure what has, or had been done with it, other than it being unboxed, and pumped down for a short period of time. Unfortunately, the people who "ran" it, are no longer affiliated with our lab. The problem, once I set the unit up and got it configured, is the autosampler itself seems incapable of doing a fast injection, despite the configuration settings in the software which have been set to tell it to. We have a similar 5972 in our lab, which I copied configuration-wise, to match the settings with that instrument. Everything else seems OK, except when it begins the injection, the plunger slowly descends down to the septa at the injection port, then there is a noticeable shudder from the tower after it injects, at which point we get an syringe error message from the tower. We are using a proper Agilent autosampler 10ul syringe as well. I've switched controller boxes with one that we know works, and also switched autosampler towers with the other 5972. Nothing is working. I'm wondering if it's something in the software, or a macro that is corrupted. Any advice will be greatly appreciated. Thanks in advance.