Dear all,

We are quite new to LC/MS (compared to GC/MS). We often use N2 generator for the drying gas. The nitrogen purity as in catalogue is about 99.5% and it's all right for us. But the flow rate is only about 9L/min. If we want to run at higher flowrate, the purity level is reduced.
I really do not know how low the purity of nitrogen for drying gas I can still run the LC/MS. Anyone have any idea about this ?

Thanks for any inputs.
The purity really doesn't have to be all that high. It's really a question of making sure that the mixture hasn't got enough oxygen to support anything explosive in the spray chamber! Generators sold for use with LC-MS equipment always have a quoted maximum flow, and you shouldn't exceed this (many generators in any case are fitted with some sort of flow-limitation device). The purity will drop with increased flow, but provided you don't exceed the maximum, it will be good enough. Most generators seem to go in steps of 15L/min, with 15L/min being sold as a "1 instrument" generator, and 30L/min designed to cope with two typical MS instruments. The exact details, though, will depend on your instruments, and whether you run them flat out at high flow-rates.
Depending on pipe-work, if you try to get too much flow out of 1 generator you may also find that the pressure drops, and your MS complains its nebuliser pressure or flow is too low.

My personal viewpoint is that if you've got a 15L/min generator and you're trying to run two 9L/min instruments, you will almost certainly find that the instruments will complain about restricted gas flow and if your generator has built-in compressors, they'll be failing frequently. You'd do better to scale up.
Thank you very much lmh. I was a bit concerned about the contamination of trace analysis if the drying gas is not pure enough. So I can increase the gas flow rate if it requires so.
3 posts Page 1 of 1

Who is online

In total there are 60 users online :: 1 registered, 0 hidden and 59 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Google [Bot] and 59 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry