Hi
Thanks for your answer
It is difficult to uderstand if the air pressure high enough.
Generator is Peak Scientific.
Pressure sensor show 0 while testing air pressure (Compessor Test Fault, all 4 compressors)
But all inside the compressor is OK, required maintenance was performed.
I have already asked authorised seller/service company and even sent a solenoid for test. But unfortunally they cannot help. And service person visit is needed indeed. We tested some by phone and e-mail, but the reason is not uderstandable. The problem is that the organization does not support this visit and.....it seems that I have 2 ways - switching to nitrogen balloongs or finding the reason of failure by myself or with volunteer help. (((((((
Looks like one of the valves is closed ans air is not comming to the sensor, but I can not test it yet(((
Air comes to the block with inscription "copmressor test", but no air input is near in the tube connected to the pressure sensor....
P.S.: sensor is OK, It shows 0 cause there is no air it the tube, when I took another tube with air and connected it to the sensor it showed pressure.
And yes...I was wrong in previous post, not nitrogen but air pressure is not enough during the compressor test.