Problems with LCMS Stability in MRM Mode
Posted: Tue Apr 05, 2005 1:05 am
have been having a lot of trouble with the stability of my mass
spectrometer (triple quad, analyzing in MRM mode).
The instrument was running very well in external and internal modes
originally, but recently runs totally unacceptable in external and
internal standard modes.
Seems to take around 5 hours for the response to stabilize. I ran
xanax/xanax-d5 as a test. After 5 hours, the external and internal
standard deviation values equalled each other. During the first 5
hours, the instrument gave much poorer external standard data.
I think it has something to do with temperature since performs much
better with all panels off. The cooling fan for the RF/DC electronics
was inoperative for about 1 month. Might have something to do with
that problem? Before the fan failed we seemed to be able to do
external calibration analyses.
I was wondering if anyone has a performance test using unlabelled
internal standards and associated statistical data. I wanted to
compare the performance of my instrument to someone elses. The
manufacturer tends to think my instrument is working properly.
I find it very unusual that the instrument repsonse take 5 hours to
equilibrate. I would expect for a well behave analysis, maybe 4-5
injections and the response should be stable.
Thanks for your help.
spectrometer (triple quad, analyzing in MRM mode).
The instrument was running very well in external and internal modes
originally, but recently runs totally unacceptable in external and
internal standard modes.
Seems to take around 5 hours for the response to stabilize. I ran
xanax/xanax-d5 as a test. After 5 hours, the external and internal
standard deviation values equalled each other. During the first 5
hours, the instrument gave much poorer external standard data.
I think it has something to do with temperature since performs much
better with all panels off. The cooling fan for the RF/DC electronics
was inoperative for about 1 month. Might have something to do with
that problem? Before the fan failed we seemed to be able to do
external calibration analyses.
I was wondering if anyone has a performance test using unlabelled
internal standards and associated statistical data. I wanted to
compare the performance of my instrument to someone elses. The
manufacturer tends to think my instrument is working properly.
I find it very unusual that the instrument repsonse take 5 hours to
equilibrate. I would expect for a well behave analysis, maybe 4-5
injections and the response should be stable.
Thanks for your help.