I have a problem with a newly developed SE-HPLC method used for stability testing: During time (0 - 3 -6 -9 months ) the target protein peak is decreasing and moving slightly to the right, also the control, a different batch of the same protein shows this behaviour. Our RP-HPLC analysis of the same protein shows no signs of decreasing signal! So it seems the problem is with the column/method.
Another tricky part is that the 5 proteins from the MW calibration marker stay relatively stabile at a constant signal level througout these different timepoints. So these 5 protein go right through the column.
The target protein is known to be sticky, but when multiple injections are made in the same sequence of samples, there is no loss in recovery! However, when the column has been stored between two timepoints the decrease in signal occurs, but ONLY for the target protein as mentioned.
Column: BioSec-3 Agilent, polymer coated silica (extra hydrophilic and pH stabile), 300 Å, 7,8 x 300mm.
Mobile Phase: PBS buffer.
Storage: after rinsing throughly with water I store the column in 20% EtOH.
Hope, you have ideas of what could be wrong!