"Below LOD"...
where LOD means "Limit of Detection"
eg: You didn't see any.
The only way to lower LOD concentration is by injecting more sample (larger volume, more concentrated prep), maximizing detector sensitivity (minimize AUFS setting, this will add some to your noise, however), or changing to a more sensitive detector (going from UV to MSD, for example), or derivitizing samples with something that a detector is more sensitive to (amino acids treated w/ OPA(?), for example).
These all require at least some linearity testing, if not revalidating the method entirely.