How to determine linearity of impurities for Isoflurane?

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
How to determine linearity of impurities for Isoflurane?
I assume you have only the isoflurane available and can separate the impurities in question.

First I would determine if the isoflurane is linear in comparison to its impurities with the detector you are using. Make serial dilutions of the isoflurane and then compare the response of the impurities to the isoflurane.

If you dilute the isoflurane and a 1% impurity is still 1% after a 1:1, a 4:1 and a 9:1 dilution (ignoring any response of the diluent) then you know the impurity responds as the isoflurane responds. If the impurity goes up or down in area % then you know you are not in a linear range for isoflurane and/or for its impurity.

This is a quick and easy experiment to perform.

Try it and let us know the results before you try making gravimetric standards.

best wishes,

Rod
Thanks a lot Rod. That helps tremendously.
3 posts Page 1 of 1

Who is online

In total there are 36 users online :: 1 registered, 0 hidden and 35 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot] and 35 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry