-There is a constant pattern to the noise from 57 to ~200 amu spaced 12-14 m/z apart and decreasing in size as the mass gets larger. Abundance of m/z57 is ~10,000 counts and m/z 200 ~500 counts.
-When I switch to EI mode (with EI source) everything appears to function normally.
-I have compared the PCI tunes to those in the past and what I can see is that the EIEnergy is dropping. It originally started out around 200 was constant for a while then dropped to ~135. Again remained constant and dropped a third time to ~77. Each drop I notice a loss of sensitivity.
Tried so far:
-Cleaned ion source
-New filament
-New HED and EM
-New methane filter and new methane from different supplier
-Cleaned transfer line and tubing
-Cleaned Quad (Agilent service)
-Changed PFDTD (Agilent service)
-Changed CI flow controller and lines (Agilent service)
Most recent tune report: Emission: 202.7, EIEnergy: 77.3, Repeller: 1.64, IonFocus: 125.3, EntLens: 15.0, EMVolts: 1106
Thanks,
Brad