I have a 5890 Series II Plus that we are trying to get running, this GC hasn't been used in years.

I have two injector ports, A and B, Injector B is running well, but Injector A is not working at all.

The pressure on injector A won't go over 0.3 psi, I checked the gas feed and no luck. I did notice that on my manual total flow control I have no gas out on the Split vent nor on the Purge Vent. I disconnected the manual flow controller and noticed that there is no gas on the outlet, replaced the controller with another used one, and same trouble, my inlet is working fine since when I unplug it and turn the gas on, there is gas coming out.

I also checked the column flow on the electronic flow module and it reads 5.35 ml/min, I tried changing it without success, I closed the gas and still have me a wrong flow reading.

Any suggestions?
could be the solenoid valve on your board.
2 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry