The request was for something else than Dionex Acclaim (Agilent in particular). Agilent does not have a WAX column as far as I know, so I suggested the equivalent. I can take both columns and by playing with amount of acetonitrle, buffer pH and buffer concentration achieve VERY similar or even the same selectivity of separation. Alternatively I can guide people to do the same.
Anyways. there are 1000s of C18 columns on the market and everybody claims "equivalency". They have diffrent loading, end-capping, pore size, pore volume, surface area, etc, but manufactureres still claiming equivalency for the purpose of method development.
According to the dictionary:
http://www.merriam-webster.com/dictionary/equivalent
Lets state the facts and you can correct me if I am wrong:
1. Both columns are based on silica gel with similar carbon load, surface area and pore size
2. Acclaim and Primesep WAX columns have carbon chain, which provide reversed-phase interaction
3. Both columns have basic group on the surface which provides anion-exchange interaction, pKa of this basic group is the same for the practical purposes of ion-exchange mechanism. pKa is about 10 and your mobile phase is usually at least 4 units away from pKa of the column.
4. Basic groups provide shielding effect of silica surface, and that is why it has less influences by source of silica
People are usually asking for equivalent columns for one of the following reason ( I don't know what was Anjie's reason):
1. Need to validate method on two different columns in order to have a back up in case something happened (very often a requirement in pharma companies)
2. Columns are not available for particular customer (delay in shipment, no stock in particular country, or problem with manufacturing)
3. Columns are expensive and less expensive alternative is required/available
4. End user had bad experience with particular company in terms of quality of the product or quality of the support (reproducibility, longevity, etc.)
5. Other reasons
So from my experience Acclaim WAX-1 and Primesep columns (reversed-phase/anion-exchange one) are equivalent to each other, and even Dionex salespeople stated about equivalency. When somebody asked for equivalent of Primesep columns. So if somebody states that Acclaim is equivalent of Primesep, so opposite should be true.
here is funny equivalency statement:
more cheese=more holes, more holes=less cheese concluding more cheese=less cheese
Lets give customers a chance to decide.
Enjoy your weekend!