Metrohm Help Needed

Posted: Sat Sep 10, 2011 1:50 am
by Jesterrose
We put some blind bids up at an auction and ended up with a few 761 Compact ICs and the associated 766 IC Sample Processors.

I am setting it up for cation analysis. I ordered the appropriate column and mobile phase reagents (nitric acid/DPA). It appears I can see ammonia.

I have 2 days to figure out how to process using IC Net 2.3. I cannot figure out how to process using duplicate injections for standards. I have to not only figure out how to process the data, I am suppose to optimize it in a production setting. Any help or ideas would be appreciated.

Re: Metrohm Help Needed

Posted: Mon Sep 12, 2011 1:53 pm
by Markus Laeubli, Metrohm
Dear Jesterrose

In ICNet you need to estabish two calibration levels for duplicate injections of standard solutions.
Each injection of any level is handled as an individual calibration point. --> 3 point calibration with duplicate injections will have 6 levels.

If you have further questions do not hesitate to contact me directly (lae@metrohm.com).