by
carls » Mon Sep 12, 2011 2:34 pm
I believe TPP is usually used as a process control, not an ISTD. The ISTD is usually something similar such as another organophosphate pesticide. If you are not looking for ethoprofos in your analysis you dont need the stable labeled ethoprofos. These methods assume the ISTD will act like the other compounds in the extraction, i.e. similar recovery, and will correct for instrument drift. However, they do not correct for matrix effects in LCMS. In pesticidce residue analysis there are typically many analytes so using stable label ISTDs for all analytes is not economically feasible.