The patient: Waters 2695, about 5 years old, under a service contract the whole time, with a single user (me) the whole time. This instrument has about 20K injections, has been regularly maintained, well treated, and stunningly reliable until last week.
The issue: Excessive but quite regular pressure fluctuations (~85 psi with a system pressure ~3K psi) when mixing gradients of methanol and water. There are no excessive fluctuations when running from a single reservoir or between two reservoirs containing like solvents. The issue ONLY happens when mixing dissimilar solvents, is worst with MeOH / water...and the piston stroke is at 100uL. It doesn't happen when the piston stroke is shorter (at a 50ul stroke, the ripple is 15-20 psi with EVERYTHING else the same).
What I've done / tested:
The first thing I did was change the filter stones on the ends of my solvent lines, thinking one or more may be slightly clogged. No improvement.
Then I tested the flow rate of the pump with methanol at ~1000psi: 1.00 mL/min set yielded 0.999 mL/min.
A static pressure test passes (accumulator decay rate = 344.3 psi/min, primary decay rate = 440.2 psi/min). A second check yielded a second passing result. No fluke there.
So, if I had to guess, I'd say that the pump, seals, check valves, and fittings are likely OK.
Next suspects: The degasser and GPV. The degasser DID have one leaky chamber which was repaired...but this had no effect on the problem (though it pumps down much more rapidly now). All others appear to be fine, and the degasser pulls down to 0.3 psia pretty quickly.
The GPV has been tested 3X and passes well within specification (error = 1.1%; specification = 3%). The line from the GPV to the inlet check valve on the primary has been replaced just as a precaution.
Just for laughs, I swapped in a priming valve from a (much) newer instrument that's running fine. Of course there was no change there...would've seen in on the static pressure test.
Our service engineer is losing his mind. He's swapped in check valves, pistons, seals, and a new primary torpedo (!!) despite passing results on tests, so those are certainly fine.
What have we missed?
(I'm kinda wondering whether the degasser is still funky, though I see no leaks and it's acting OK).