Metrohm Help Needed

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
We put some blind bids up at an auction and ended up with a few 761 Compact ICs and the associated 766 IC Sample Processors.

I am setting it up for cation analysis. I ordered the appropriate column and mobile phase reagents (nitric acid/DPA). It appears I can see ammonia.

I have 2 days to figure out how to process using IC Net 2.3. I cannot figure out how to process using duplicate injections for standards. I have to not only figure out how to process the data, I am suppose to optimize it in a production setting. Any help or ideas would be appreciated.
Dear Jesterrose

In ICNet you need to estabish two calibration levels for duplicate injections of standard solutions.
Each injection of any level is handled as an individual calibration point. --> 3 point calibration with duplicate injections will have 6 levels.

If you have further questions do not hesitate to contact me directly (lae@metrohm.com).
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland
2 posts Page 1 of 1

Who is online

In total there are 13 users online :: 2 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 11 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry