Page 1 of 1

Full Reporting In empower

Posted: Wed Sep 07, 2011 4:02 pm
by cody84
Hey I was wondering if anyone know how to create a FULL report in Empower no matter how many samples/injections are run.

Here is what would be ideal:
Page 1: System suit result, 6 chroms overlayed and offset, summary table showing related parameters, possibly a PASS/FAIL result.
Page 2: blank injection chroms, overlayed
Page 3: First two bracketing standard injections overlayed and summary table (2 injections total)
Page 4: First sample vial with duplicate injection overlayed on small chrom, second sample vial with duplicate injection overlayed, summary of each vial and then summary of both vials results
Page 5: Next set of bracketing standards
Page 6: Next sample etc.

Here is my problem:
So far I can have everything work perfectly except that ALL of the standards show up on one chromatogram and page. Is there any labelling solution to this that is open ended and won't be affected by how many standards are injected in the sample set?

thanks in advance for any help :)

Re: Full Reporting In empower

Posted: Wed Sep 07, 2011 5:21 pm
by DR
The only way I can think of is to make copies of the standard info and then apply appropriate filters to the "order by" tab on each copy.

Let's assume your first set of standard injections is labeled S0101, S0102... 2d set is S0201, S0202... and so on.

1st filter would be for label S01* and would yield info for every standard in 1st group. Filter for 2d copy would be S02* and so on. Make as many copies as needed knowing that you will get a "no data for... error at the end of the report if you have more of these than you have sets of standards (not a major concern around here).

If there is a better way to do this, one of our resident experts will surely clue us in.