by
Hollow » Mon Sep 05, 2011 12:03 pm
as others, my first thought would be too strong on sample diluent.
Will the peak split gets better if the injection volume is reduced?
->try a injection linearity e.g. 5, 3, 2, 1 µL and see if the response (Area/Vinj) is stais the same or if there is a deviation from linearity.
If you have a PDA detector than you probably compare the spectra of the two peaks (hopefully the impurity has some different sepctra)
If you're using the same model of HPLC in the two labs, then are the diameter of the capillary (especiall inj - column) also the same or is QC using smaller id capillaries (on all of their systems). Where they changed since validation?
-> smaller capillary would make the system more sensitive to strong samples solvents effect.
Another possibility would be void volumes somewhere in the systems of QC.
Where they recently maintained? ->check, re-fit every connection.
Are they using adjustable fittings and do the installation of the column properly?
Or do they have fixed fittings that were set for some other column type (eg. Waters vs. Agilent columns have different depths of the connection)?