how does retention time change from 30m to 60 m long DB5 column? I usually use solvent delay of 5 minute in the 30 m long column. What should be the corresponding solvent delay in 60 m long. Can we relate with the K value only? I used twice the time and 1.5X the column flow but did not work. Does anyone know what should be the proper setting? posting.php?mode=post&f=2#
-
- Posts: 377
- Joined: Wed Aug 24, 2011 2:10 pm
Why don't you just re-check the solvent retentio time? Way I used to do this was to use an empty vial and with the autosampler wash solution the same as your analysis solution. You get a small wash solvent carryover peak, small enough for the MS to handle, and you can then set your solvent delay accordingly.
Where can I buy the kit they use in CSI?
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
Depending on what you are doing, you do not need to worry about when the solvent elutes, simply turn the filament on before the first peak of interest elutes.
Peter
Peter
Peter Apps
-
- Posts: 102
- Joined: Mon Jan 10, 2011 1:38 pm
if you translate your present method to a longer column the right way, you will have exactly double the retention time. In order to get the same chromatography, you need to adjust your oven temperature program also, so you get the same elution temperatures.
The average linear gas velocity should stay the same. For this you need probably to double the inlet pressure. solvent delay time should also be twice as long.
Concerning your program conditions, roughly iso times must be doubled and program rates must be Halved.
Jaap de Zeeuw, Restek corporation
The average linear gas velocity should stay the same. For this you need probably to double the inlet pressure. solvent delay time should also be twice as long.
Concerning your program conditions, roughly iso times must be doubled and program rates must be Halved.
Jaap de Zeeuw, Restek corporation
-
- Posts: 11
- Joined: Tue Sep 06, 2011 4:18 pm
Thank you all for valuable comments.
5 posts
Page 1 of 1
Who is online
In total there are 17 users online :: 0 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 17 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 17 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science