Square solvent peak!

Posted: Sun Apr 03, 2005 8:06 pm
by rakela
Help!
I'm using a GC-FPD for organotin analysis and am getting a large square solvent peak at the outset, which can be very large and produce ugly chromatograms (although it doesn't interfere with my analysis as such). My solvent is iso-octane... Any suggestions?
Rachel :)

Posted: Mon Apr 04, 2005 1:24 pm
by Answer Man
Unless the solvent peak is causing quantitation problems, or extinguishing the flame, I would not be concerned - it is common that the solvent overloads the column and detector, which causes a flat-topped peak.

You may find the following link helpful.

http://www.restek.com/fantasia/pdfCache/59550.pdf

Posted: Mon Apr 04, 2005 3:36 pm
by DR
If it is significantly larger than what you're used to getting - check for leaks or misprogramming of the injector.

Re: Square solvent peak!

Posted: Tue Apr 05, 2005 7:29 am
by xxx
Help!
I'm using a GC-FPD for organotin analysis and am getting a large square solvent peak at the outset, which can be very large and produce ugly chromatograms (although it doesn't interfere with my analysis as such). My solvent is iso-octane... Any suggestions?
Rachel :)
:D I have the same thing when i analyse CS2 in isooctan,the solvent peak at the end of the chromatogramm is very large...but don't interfere...