Removal Of Trace Oxygen From HPLC Mobile Phase
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
3 posts
Page 1 of 1
We are needing a HPLC %area purity method on a VERY oxygen sensitive compound. We have premixed and sonic de-gassed our phases (MeCN & Water), and are using an in-line vacuum degasser. However, we are still seeing significant oxidation of our product. Does anyone have any experience of removing even more oxygen from the mobile phase?
-
- Posts: 31
- Joined: Mon Jul 19, 2010 11:18 pm
Hi,
You can probe using a vacuum system as same time you are sonicating the mobile phase. It is very effective.
Regards
You can probe using a vacuum system as same time you are sonicating the mobile phase. It is very effective.
Regards
-
- Posts: 2174
- Joined: Tue Aug 17, 2004 7:59 pm
Or just displace the oxygen w/ helium (continuous low flow He sparge after initial degassing).
Thanks,
DR

DR

3 posts
Page 1 of 1
Who is online
In total there are 364 users online :: 2 registered, 0 hidden and 362 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot], Google Adsense [Bot] and 362 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot], Google Adsense [Bot] and 362 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science