a negative peak in anion determination by ion chromatography
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
6 posts
Page 1 of 1
The method that I am using for phosphite and phosphate determination is ion chromatographic. The problem is that during the measurement a negative peak appeares with moving retention time, even in the blank sample which is water. What can be the reson?
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
It's called a "system peak". It's caused by the disturbance in equilibrium due to the injection process. Not at all unusual to see it with a conductivity detector.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 3
- Joined: Wed Aug 24, 2011 12:52 pm
I don`t think that this is a "system peak" because the retention time of this peak is moving to more negative values when one and the same sample is injected. And this peak appeared only in the last analysis but not in the previos ones.
-
- Posts: 362
- Joined: Tue Aug 31, 2004 6:54 am
Please give more details e.g. column, eluent ...
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland
-
- Posts: 3
- Joined: Wed Aug 24, 2011 12:52 pm
We found out the reson. This peak comes from the water, it wasn`t pure enough.
-
- Posts: 48
- Joined: Thu Apr 14, 2005 10:26 am
Typicaly, negative peaks in ion chromatography (and other modes as well) come from some eluent constituent or contamination, which is giving response on the detector. In case the concentration of such compound in sample is lower than in eluent, you are getting the negative peak. The reason for the retention time shift may be in column or precolumn saturation (some sort of frontal chromatography may take place in case the contaminant has high affinity to the column)
6 posts
Page 1 of 1
Who is online
In total there are 317 users online :: 1 registered, 0 hidden and 316 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 316 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 316 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science