can I use a bog standard syringe filter to get rid of salts, or do I have to go down the fancy SPE road?
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
I'm not sure I understand the question
. More detail on how were you proposing to desalt using a filter would help.

-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 252
- Joined: Sat Nov 07, 2009 6:27 pm
golly, I think this is the first time i've heard "SPE" and "fancy" together in one sentence.
spe is cheap, effective- available in 96 well format. Unless sample loss is a concern... Are you talking about something akin to membrane dialysis, fit on a syringe?

spe is cheap, effective- available in 96 well format. Unless sample loss is a concern... Are you talking about something akin to membrane dialysis, fit on a syringe?
-
- Posts: 2846
- Joined: Mon Aug 30, 2004 7:17 am
This is usually done with cenrifuge ultrafilters, dialysis chambers, ion exchange, or size exclusion.
-
- Posts: 65
- Joined: Fri Jul 01, 2011 1:29 pm
First of all i like your nick.... Second: What exactly do you want to desalt? If you want to desalt your mobile phase before entering into MS, then you chould follow this link.http://www.dionex.com/en-us/products/ac ... 73569.html. The so called suppressor is able to desalt and neutralize the mobile phase by converting the salt into the acid. Your friendly Dionex/ThermoFisher rep would be happy to help you in this case.
Stefan Brand
Thermo Fisher Scientific
Thermo Fisher Scientific
-
- Posts: 264
- Joined: Sun Nov 14, 2010 2:20 pm
Do you really need to desalt before before RP LC/MS? Would not he salts come out at the void or very close to it in most cases... Could not you just divert the flow to waste until after the salts elute?
- Karen
- Karen
-
- Posts: 129
- Joined: Wed Apr 01, 2009 4:02 pm
@Karen01 - That what I did. I had a guard column before the column;then after another switch valve that just sent the first part of the liquid to waste and then back online to the MS.
7 posts
Page 1 of 1
Who is online
In total there are 345 users online :: 0 registered, 0 hidden and 345 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 345 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 345 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science