Increasing background, poor standard reproducibility
Posted: Thu Aug 25, 2011 5:27 pm
Currently we have been analyzing a specific set of PCB congeners in biological samples on our Agilent 5975C/7890A GC-MS. During our method development stage, we have been finding that when we run 1 set of standards, our extracted samples and then re-run a set of standards, the peak areas for the second set of standards (all though they are the exact same batch as the first) increase by as great as 50% along with the baseline, which slowly goes back to the initial value after running a number of standards. Our recoveries for spiked samples are almost always greater then 100%. However when we complete repeat injections for the same standard, peak areas are approximately the same (%RSD <5)
We have tried the following:
Clean inlet, change liner types (shape, w and w/out glass wool), different syringes, re-installing column after trimming about 15 cm, using internal standards, changing injection mode (splitless, pulsed splitless etc).
The MS has been tuning fine and we do not suspect that is where the problem lies.
Also we have ran the same extraction procedure and analyzed by uECD and the results were very reproducible, therefore we don't think it is sample related
GC MS method is: 2 uL pulsed splitless injection inlet liner 275C. Column: 30m DB-5ms 0.25x250. Inital temp of 100C ramp to 180 @25; ramp to 250 at 3; ramp to 280@20 hold 5 mins. Flow is 1 mL/min. MS is EI in SIM mode with gain factor of 5. Source temp of 250, Quad 150
Extraction procedure is:
Vortex with 1:1 acetone:hexane 3x. concentrate to 1mL; elute from 15 mL florisil SPE cartridge, concentration in multivap to ~200 uL.
Final extract is hexane, all standards prepared in hexane (from stocks in isooctane)
Attached is a chromatogram with the 0.5 ppb standard before (Black) and after(Blue) the extract injections
Thank you for your help