Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
We are performing traditional gradient elution and UPLC-QQQ-MS for some pharmaceuticals with 2 mmol ammonium formate, 0.05% formic acid buffer (A) and acetonitrile (B). Retention times are stable for the four lowest concentration levels, but there are significant retention time changes for some compounds in higher concentrations. Run-to-run variations injected from the same vial are negligible (so the problem is concentration dependent). Column is typical C18 2.1 x 50 mm, 1.7 µm.

Is this all about too weak buffering or could there be some other reasons?
very interesting...

What is your concentration range?
Are significant retention time changes in higher concentrations for both standard solutions and real samples or just for real samples? How much do they differ? Is there any splitting?
What is the injection volume?
What is peak asymmetry at 1/10 of height for high concentrations?
2 posts Page 1 of 1

Who is online

In total there are 67 users online :: 2 registered, 0 hidden and 65 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Bing [Bot], Google [Bot] and 65 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry