Ethanol carryover problem with headspace
Posted: Wed Aug 24, 2011 2:59 am
We do an annual calibration on our GC's at work and most have a headspace connected to them. We have Agilent 6890's with mostly 7694 headspaces (a couple G1888). We use Agilents protocol for the headspace: 5m x 0.1mm column with no phase, 200ºC isothermal oven, small glass capillary tubes that you angle in ethanol to suck it up and placed in a 20mL headspace vial. Headspace is set for 90ºC oven, 110ºC loop and 115ºC transfer line. The headspace used in this discussion is an older Agilent 7694.
Well, a fellow chemist said that a former chemist used to just put a few drops of Ethanol in the vials and not bother with the capillary tubes. I decided to give it a shot and did two test injections. The peaks were enormous and fat and about 5x larger than what we typically get so I decided it wasn't a good idea. I prepped the 6 test vials followed by a blank (to measure carryover, which must be < 0.1% to pass) and ran them. I got 4.2% RSD (limit is 2%) and 14.8% carryover! I then ran another blank and the carryover was smaller but not much. I then had an idea...
I set the GC to run for 15 minutes and I went into manual control on the headspace. Every 10 seconds I would open the sample valve, which puts the carrier flow through the sample loop, the transfer line and into the GC inlet. Well, every 10 seconds I get a very large peak, almost as big as the one in my carryover injection. I repeated this about 20 times; open the valve for 10 seconds, close it for 10 seconds (to allow the ethanol to accumulate) and each time I get a peak that is still huge but decreasing in size. Even when I did it every 5 seconds I would get a large peak. It's like the ethanol is endless!
This is the first time I've ran into a carryover problem so I decided to run Agilent's steam cleaning procedure, which has you inject 20 water vials (1mL water into 20mL vial) to clean out the flow path. After cleaning I ran 6 new ethanol vials and a carryover. I got 3.2% RSD this time and a much better 4.6% carryover. I ran the steam cleaning again! This time I got 1.5% RSD and 3.2% carryover. Better, but still 32x the allowed carryover!
We run a lot of methods that use DMSO and DMA, if that matters. I can't believe that me running the highly concentrated ethanol test vials in the beginning would have caused this but who knows. Does this sound like a contaminated loop or transfer line? Is there any way to decontaminate/deactivate it? Just replace?
RECAP:
First run after high concentrated test injections: 4.2% RSD 14.8% Carryover
After first steam cleaning: 3.2% RSD 4.6% Carryover
After second steam cleaning: 1.5% RSD 3.2% Carryover
Thanks in advance for any help! It's greatly appreciated!
Well, a fellow chemist said that a former chemist used to just put a few drops of Ethanol in the vials and not bother with the capillary tubes. I decided to give it a shot and did two test injections. The peaks were enormous and fat and about 5x larger than what we typically get so I decided it wasn't a good idea. I prepped the 6 test vials followed by a blank (to measure carryover, which must be < 0.1% to pass) and ran them. I got 4.2% RSD (limit is 2%) and 14.8% carryover! I then ran another blank and the carryover was smaller but not much. I then had an idea...
I set the GC to run for 15 minutes and I went into manual control on the headspace. Every 10 seconds I would open the sample valve, which puts the carrier flow through the sample loop, the transfer line and into the GC inlet. Well, every 10 seconds I get a very large peak, almost as big as the one in my carryover injection. I repeated this about 20 times; open the valve for 10 seconds, close it for 10 seconds (to allow the ethanol to accumulate) and each time I get a peak that is still huge but decreasing in size. Even when I did it every 5 seconds I would get a large peak. It's like the ethanol is endless!
This is the first time I've ran into a carryover problem so I decided to run Agilent's steam cleaning procedure, which has you inject 20 water vials (1mL water into 20mL vial) to clean out the flow path. After cleaning I ran 6 new ethanol vials and a carryover. I got 3.2% RSD this time and a much better 4.6% carryover. I ran the steam cleaning again! This time I got 1.5% RSD and 3.2% carryover. Better, but still 32x the allowed carryover!
We run a lot of methods that use DMSO and DMA, if that matters. I can't believe that me running the highly concentrated ethanol test vials in the beginning would have caused this but who knows. Does this sound like a contaminated loop or transfer line? Is there any way to decontaminate/deactivate it? Just replace?
RECAP:
First run after high concentrated test injections: 4.2% RSD 14.8% Carryover
After first steam cleaning: 3.2% RSD 4.6% Carryover
After second steam cleaning: 1.5% RSD 3.2% Carryover
Thanks in advance for any help! It's greatly appreciated!