Analysis of Hexaconazole

Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
Anyone of you has idea to analyses this pesticide compound? As i know it is determined by GC-NPD, but my lab just have GC-FID only.

Thanks.
Hi

GC-FID works, its just not selective like the NPD. As long as its has the sensitivity and you can separate the hexaconazole from anything else eluting you will be fine.
Just use a base deactivated liner to help reduce tailing.
If I'm not mistaken, boiling point for Hexaconazole is around 490 degree celcius. My supplier told me that most of the FID detector couldn't stand at that high temperature? is it true?
Depending upon the GC you're using, set the base of the detector to 340. (For H-P it need to be higher). Hexaconazole is very difficult by GC - you're way better off using LC for this compound.
Mark Krause
Laboratory Director
Krause Analytical
Austin, TX USA
4 posts Page 1 of 1

Who is online

In total there are 191 users online :: 1 registered, 0 hidden and 190 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Bing [Bot] and 190 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry