I'm currently running my drug of abuse project on abs 3200 qtrap Lcms/ms.. However,recently I found that after few days of my run, the signal will became unstable, and the vacuum gauge reading will drop.. I suspect that this is due to contamination/blockage in the orifice plate. And as each time it happened, I wud clean the orifice, and find that the signal will turn alright again and the vacuum gauge reading will increase back to normal.
What has really been bugging me, is that this problem has been occurring for few times, in about 3 weeks time...which is not normal, as this problem shud not happened so frequent.
I check for the instrument status, on each of my run.. And it turns out to be okay,: the operating temperature reached, and everything else is normal..except for the vacuum gauge reading, which will drop.. I set the temp to 550 degrees, and both gas 1 and gas2 at 40..
I doubt that it is due to improper sample cleanup, as I do spe to extract my samples..
Has this problem ever occurred to anyone before?