Dimethylamine hydrochloride tests
Posted: Fri Aug 19, 2011 5:18 pm
by vishnu
Tests either in hplc or in gc needed. is there any method for testing dma hcl (N-methylmethamine) by gc or hplc???
please help.
Re: Dimethylamine hydrochloride tests
Posted: Mon Aug 22, 2011 1:00 pm
by Markus Laeubli, Metrohm
I would recommend cation exchange chromatograph (IC) with conductivity detection.
See a typical column out of the Metrohm range:
Metrosep C 4 - 140/4.0
Eluent conditions: 1.7 mmol/L nitric acid, 0.15 mmol/L dipicolinic acid, 3% acetonitrile (0.9 mL/min).
Posted: Mon Aug 22, 2011 3:25 pm
by pharmaguy
Fastest way is probably GC - I seem to recall some even sell "amine" columns.
Posted: Mon Aug 22, 2011 4:23 pm
by vishnu
Thanks for all for the reply.
I think there is a misunderstanding in my words.
I will clarify.
I am asking about DIMETHYLAMINE HYDROCHLORIDE
All the suggestions above are for DMA.
Posted: Mon Aug 22, 2011 8:50 pm
by tom jupille
In aqueous solution, they are the same thing: dimethylammonium ion.
Posted: Tue Aug 23, 2011 1:28 am
by vishnu
Thanks for the clarification.
Is there any method using ODS column (C-18) or a cationic SCX column???
Bcoz i have these both columns in hand. Can i test using these columns???
If it is there, please explain me the method.
Posted: Tue Aug 23, 2011 5:05 am
The scx column (cation exchange) might work.
Us an acidic eluent with some organic modifier with a conductivity detector.
The dimethylammonium cation is retained on the cation exchanger material and will be eluted mainly by the H+ of the eluent acid. The organic modifier speeds up the separation and improves the peakshape.
Posted: Tue Aug 23, 2011 6:27 am
by DSP007
Hello Vishnu.
Use
GC wich Fid ; packing column (5% Carbowax+ 1% KOH * on S1), head space (90 C) wich water solution or extract on CHCL3 + one drop of NH3 solution.
inj - 100c det -150C , column 60->120 5 C/min.
You detremine [url]dimethylamine as base[/url] and calculate by hydrochloride.
Ion chromatography (HPLC) admixting iones in drug substancies
in a salt forms - is a big problem...
Posted: Tue Aug 23, 2011 7:41 am
by vishnu
Is UV visible detector effective for that in HPLC???
At what nm i can use for DMA???
Posted: Wed Aug 24, 2011 5:11 pm
by Vlad Orlovsky
you cannot use UV for this compound, despite the fact that some people think. Your choices for this are LC/MS, ELSD, RI, conductivity and nitrogen detector. if you want to see both ions 9amine and chloride) you need trimodal column like Obelisc N or Obelisc R. If you are looking only at amine you can use Primsep mixed-mode column, similar to this:
http://www.sielc.com/Compound-Ethylamine.html
you can use your SCX column in the same fashion with ELSD, RI or conductivity
Posted: Thu Aug 25, 2011 8:41 am
by DSP007
Dimethylamine do'n have chromoforic groups (double bonds).
You can (should) use a admixting of chromofors (benzene , nafthalene and any dues) in liquide phase and register negative peaks in UV-DAD , but this LC- method is'n effective, selective and robatic.
I say. Dear Vishnu "do not torture yourself and chemin de fer
" - use a valid GC method by packing column...
Posted: Thu Jan 31, 2013 11:56 pm
by 5.56x45mm
your Ca and Mg peaks are miss identified. They should be reversed, with Ca eluting before Mg.
http://www.metrohmusa.com/Products/Ion- ... Cbr+%2F%3E
Posted: Fri Feb 01, 2013 8:55 am
by HPLCaddict
please help.
I think the most important question hasn't been asked yet
. WHAT kind of test do you exactly need?
Verifying identity? Assay? Purity? Anything else?
Nevertheless I'd say my choice would be GC...
Posted: Fri Feb 01, 2013 3:30 pm
by HPLCaddict
Oops, just realized that this thread is more than a year old