5890 II plus inlet temperature fluctuating

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
I had problem with the GC 5890 II plus, the inlet temperature is not stable, It is actuating by 5+/- from the set point, I did replace the heater/sensor assembly and also the main-board and also I did put more insulation in the inlet port with out any success. Do you guys have and idea about what is going on.
You need to provide more information:

Setpoint of injector
Oven temperature, including program profile

Thanks
You need to provide more information:

Setpoint of injector
Oven temperature, including program profile

Thanks
Thanks I got it fixed. It was the heater sensor assembly that i put new.