5890 II plus inlet temperature fluctuating
Discussions about GC and other "gas phase" separation techniques.
3 posts
Page 1 of 1
I had problem with the GC 5890 II plus, the inlet temperature is not stable, It is actuating by 5+/- from the set point, I did replace the heater/sensor assembly and also the main-board and also I did put more insulation in the inlet port with out any success. Do you guys have and idea about what is going on.
-
- Posts: 7
- Joined: Sat Feb 26, 2005 1:25 am
You need to provide more information:
Setpoint of injector
Oven temperature, including program profile
Thanks
Setpoint of injector
Oven temperature, including program profile
Thanks
-
- Posts: 80
- Joined: Thu Feb 11, 2010 10:58 pm
Thanks I got it fixed. It was the heater sensor assembly that i put new.You need to provide more information:
Setpoint of injector
Oven temperature, including program profile
Thanks
3 posts
Page 1 of 1
Who is online
In total there are 9 users online :: 1 registered, 0 hidden and 8 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Bing [Bot] and 8 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Bing [Bot] and 8 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science