Linearity Problem GC Autosampler
Posted: Fri Aug 19, 2011 4:36 am
My dear colleagues ;
I have a linearity problem with a autosampler GC method and its story is stated below ;
I have been asked to make partial gc method validation of a drug master file method which contains solvents as petrol ether(70-90°C), acetone, methanol,dichloromethane,acetonitrile. Methold conditions are listed below ;
GC HP 7890
Column : hp-wax 30mx0.53mm 1µl ( I used varian cp wax 52 cb and restek mxt-wax with same dimensions)
diluent : NNdimethylacetamide
GC autosampler
inlet temp : 140°C
1 µl injection volume
3 psi constant flow
splitless
oven program : 40°C 20 min 6°C/min 165°C 20 min
sample con. : 20mg/ml
solvents limits : petrol ether(70-90°C)50ppm, acetone 250ppm, methanol 300ppm,dichloromethane 300ppm,acetonitrile 200ppm
Due the bad peak shapes I made little modifications like
4:1 splıt , 45°C starting temp. and 3,6 psi constant flow
When I had started method validation a unexpected problem occured. Petrol ether and acetone were linear but rest solvents are nonlinear. Their linearity curve is more like Nike sports wear logo
a little wave.
No matter how many times I trıed (different splıt ratios even injection volumes) again and again results did not change.
But when I increase the sample concentration like 2 fold and respectively solvent concentration and if I make analysis splıtless with awful peak tailing ,suprisingly they were all linear,
I have tried 2 different liner types, 3 different GCs, 2 different GC columns which were brand new but resuts are same.
Any comments will be highly appreciated.
Best Regards.
I have a linearity problem with a autosampler GC method and its story is stated below ;
I have been asked to make partial gc method validation of a drug master file method which contains solvents as petrol ether(70-90°C), acetone, methanol,dichloromethane,acetonitrile. Methold conditions are listed below ;
GC HP 7890
Column : hp-wax 30mx0.53mm 1µl ( I used varian cp wax 52 cb and restek mxt-wax with same dimensions)
diluent : NNdimethylacetamide
GC autosampler
inlet temp : 140°C
1 µl injection volume
3 psi constant flow
splitless
oven program : 40°C 20 min 6°C/min 165°C 20 min
sample con. : 20mg/ml
solvents limits : petrol ether(70-90°C)50ppm, acetone 250ppm, methanol 300ppm,dichloromethane 300ppm,acetonitrile 200ppm
Due the bad peak shapes I made little modifications like
4:1 splıt , 45°C starting temp. and 3,6 psi constant flow
When I had started method validation a unexpected problem occured. Petrol ether and acetone were linear but rest solvents are nonlinear. Their linearity curve is more like Nike sports wear logo
No matter how many times I trıed (different splıt ratios even injection volumes) again and again results did not change.
But when I increase the sample concentration like 2 fold and respectively solvent concentration and if I make analysis splıtless with awful peak tailing ,suprisingly they were all linear,
I have tried 2 different liner types, 3 different GCs, 2 different GC columns which were brand new but resuts are same.
Any comments will be highly appreciated.