100 mMol mineral acid for RP column?

Posted: Mon Aug 15, 2011 11:54 pm
by sepscientologist
Bischoff shows a beautiful separation of organic acids on a carbamate phase with
100 mMolar phosphoric acid as a mobile phase. Has anyone tried something like
that with a reversed phase column? The pH must be around 1. I would like to try
it but I'm afraid I'll lose my stationary phase before I even equilibrate.

Re: 100 mMol mineral acid for RP column?

Posted: Tue Aug 16, 2011 1:14 am
by tom jupille
There *are* columns with special bonding chemistry that will survive down that low (for example, the "sterically protected" phases in the Zorbax StableBond) but they are the exception rather than the rule. I would check with the manufacturer of the specific column you plan to use and establish the recommended pH range.