Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
7 posts
Page 1 of 1
Waters HPLC Alliance shows solvent ratios on monitor screen as 100%A, 0%B, 0%C, 0%D Empower on computer shows 50%A, 0%B, 0%C, 0%D. Sample sets will not run with discrepancy. Should I re-config instrument on software? How can it only add up to 50% on software???
-
- Posts: 2175
- Joined: Tue Aug 17, 2004 7:59 pm
Are you initiating running via Empower, or by using the keypad on the Alliance?
Thanks,
DR

DR

-
- Posts: 16
- Joined: Tue Jun 14, 2011 12:41 pm
Initialization is through the Empower software.
-
- Posts: 2175
- Joined: Tue Aug 17, 2004 7:59 pm
So you tell it to pump x mL/min, 50%A, 50%B and the Alliance does so and reports doing so on the Alliance display but is fouled up in Empower...
I'm thinking that there may be a firmware issue with the Alliance. Have you recently installed anything new on the Empower server(s)?
If it was working fine until recently, and there have been no changes to the LC or Empower servers, I'd go through the cable connections very carefully, making sure they're all snug and clean.
I'm thinking that there may be a firmware issue with the Alliance. Have you recently installed anything new on the Empower server(s)?
If it was working fine until recently, and there have been no changes to the LC or Empower servers, I'd go through the cable connections very carefully, making sure they're all snug and clean.
Thanks,
DR

DR

-
- Posts: 16
- Joined: Tue Jun 14, 2011 12:41 pm
OK I will check connections. When i tell Empower to set at 100%A it goes 100% A on alliance but 50%A on Empower.
When I tell it 75%A and 25%B in empower, it goes to 37.5%A and 12.5%B (exactly half)
No updates have been done recently to firmware or system. strange....
When I tell it 75%A and 25%B in empower, it goes to 37.5%A and 12.5%B (exactly half)
No updates have been done recently to firmware or system. strange....
-
- Posts: 2175
- Joined: Tue Aug 17, 2004 7:59 pm
If checking cables is no help, try writing a new instrument method from scratch and see if that works.
If not, please post a lot more detail about your Empower setup and Alliance.
If not, please post a lot more detail about your Empower setup and Alliance.
Thanks,
DR

DR

-
- Posts: 16
- Joined: Tue Jun 14, 2011 12:41 pm
Thanks for your help. I took the instrument offline and then online again. I re-booted the alliance and now it works fine! It must be a bug when they updated the server.
7 posts
Page 1 of 1
Who is online
In total there are 44 users online :: 2 registered, 0 hidden and 42 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Baidu [Spider], Google [Bot] and 42 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Baidu [Spider], Google [Bot] and 42 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science