Shifting Standard Peaks (GPC)
Posted: Mon Aug 15, 2011 9:06 am
Hello all,
I have spent the last few days trying to solve a number of (possibly unrelated) problems with our GPC [Agilent 1200 Series pump and autosampler, Phenomenex column, Wyatt detectors]. Our Wyatt engineer is satisfied that the detectors and analysis software are working well, and the chromatography obtained looks clean, with good peak shapes.
However, the peaks themselves are not at the expected retention time. For example, peaks for a BSA standard (MW ~70 kDa) have been seen before eluting at around 8 minutes. I am know seeing them at around 11 minutes. Similarly, the main polymer with which our research group works (MW ~10-12 kDa), which will normally elute at around 11 minutes, isn't being seen at all; the injections of this polymer consistently cause a visible disturbance to the injection (solvent) peak at around 16 minutes.
If the pores on the GPC column were becoming blocked, I would expect my peaks to shift earlier rather than later. Given that the reverse is true, the best I can hypothesise is that there is some problem with the autosampler/injection mechanism.
Any advice or suggestions would be very gratefully received - I'm fast approaching the end of my tether on this one!
Many thanks,
Chris.