100 mMol mineral acid for RP column?

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Bischoff shows a beautiful separation of organic acids on a carbamate phase with
100 mMolar phosphoric acid as a mobile phase. Has anyone tried something like
that with a reversed phase column? The pH must be around 1. I would like to try
it but I'm afraid I'll lose my stationary phase before I even equilibrate.
There *are* columns with special bonding chemistry that will survive down that low (for example, the "sterically protected" phases in the Zorbax StableBond) but they are the exception rather than the rule. I would check with the manufacturer of the specific column you plan to use and establish the recommended pH range.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 409 users online :: 1 registered, 0 hidden and 408 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot] and 408 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry