Which requirements are requested for %RSD of peak areas for replicate injections for impurities testing.
Our internal specification is for 5 replicate injections of a reference solution, %RSD < 5 %.
But I don't find a justification (requested by the authorities). Do You have any suggestions.
Dear Majorel,
For impurity determination methods (RS), generally Resolution will be sufficient as a system suitability to check the column performance. However, to check the precision, replicate injections of the diluted standard solution is good enough. Keeping the diluted standard concentration of 10 ppm or less, % RSD of 5 is acceptable. Many of the USP monographs including Cefpodoxime proxetil will be good reference to you to justify.
Good luck,
Warm Regards
Murugan Saravanan
about analitical determination of samples (non impurities) wich %RDS is recomended?2 %? or less...
thanks
Whatever is appropriate to the manufacturing tolerance of the product. If the content of the product must be controlled within a +/-5% window (for example), then you need to measure with 2% RSD (approximately).
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
is the gold metrology rule? in this case correspond the precision must be 5 to 10 more of the request precision, for example, for specification 90% to 110% of labeled amount, a 1 to 2 %RDS would be aceptable
Thanks
Depends on the confidence level required! This is the sort of thing where you want to have a statistician look at what you are trying to do.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Is there any good reference for this info? For example here QA considers all all methods should have <2% RSD, 1.5 resolution, peak tailing < 2.0, recovery of 98 - 102 (for ALL test methods) and so on. Is there any really good references I could cite for their satisfaction? Otherwise we are stuck with these insane specifications that are set out in an SOP (even for trace metals in ppb).
I once read somewhere that the expected %w/w of whatever you are testing is what determines these parameters. And from reading abstracts and journals, I always see results that are way outside of what would be acceptable here and yet they say "the method was found to be valid blah blah" :) Any help would be greatly appreciated! Thanks guys!
The starting point is the allowed variation in your process or product (if your QA people don't understand *that*, they are in the wrong line of work! :wink: ). That will determine the amount of uncertainty you can have in your measurement. There was a brief discussion early last year:
viewtopic.php?f=1&t=12109

Once you have the uncertainty, you have to define the required confidence interval (usually 95% or 99%). Those, in turn, will set the allowed recovery range (which will depend on the number of replicates!).

Validation is defined as "demonstrating that the method does what it purports to do".

One of the parameters to be determined is robustness: how gracefully the method tolerates variations in conditions. The robustness data are what allow you to set reasonable system suitability parameters.

If your robustness data show that you can attain acceptable accuracy and precision with Rs = 1.1, but not when Rs drops to 1.0, then you are justified in setting Rs = 1.1 as a system suitability parameter. That said, it would be hard to do in practice. Rs = 1.5 basically provides 1% overlap of two identical symmetrical peaks. The US FDA in their "Guidance" *suggest* that Rs should be > 2; that is *not* a requirement, merely a suggestion. In practice, you can get away with less in some cases and more often require substantially more (e.g., in quantitating trace impurities, or if significant tailing is present).

A parallel comment applies to tailing. The FDA suggests that TF should be < 2 (they seem to have a fixation on the number "2" :wink: ). As with Rs, you should set system suitability based on what your robustness data justifies. If you can demonstrate acceptable quantitation with TF = 2.5 but not with TF = 2.8, then set TF <= 2.5 as system suit. In practice, quantitation starts to degrade when tailing factor exceeds 1.5, and most chromatographers would consider 2.0 to be generous.

In our Advanced HPLC Method Development course, I suggest to attendees that they add whatever other criteria can be useful in avoiding bogus results. Things like baseline noise (for trace analysis), retention time (or better, k') windows, and so on.

Whatever you do, it must be based on good science. If the SOP is *not* based on good science, it should be modified.
<2% RSD, 1.5 resolution, peak tailing < 2.0, recovery of 98 - 102
are fairly typical, but they may be gross overkill in some situations and grossly inadequate in others. The flaw is assuming that "one size fits all".
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Thanks!
So when and where are you teaching the course? :)
I'm glad you asked! :wink:

The schedule is on our web site:
http://www.lcresources.com/training/coursedate.htm
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
10 posts Page 1 of 1

Who is online

In total there are 379 users online :: 0 registered, 0 hidden and 379 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 379 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry