Agilent 6890/Autospec
Here's what I've done.
Finished one project.
Remove column from that project.
Replace GC liner, injector septa, gold plated injector seal.
Install column1
Run samples, all is good, including last check standard after run (valley between co-eluting peaks = 6%).
No one else uses this instrument and it sat over the weekend as routinely does.
Run standards, poor peak shape. (no instrument/column changes made between "good" and "bad" runs) (valley between co-eluting peaks = 15%)
Replace retention gap (deactivated FS, ~1M), Union and 1 loop of column1 (DB-5MS, 15M, remove ~0.5M) (valley between co-eluting peaks = 18%, no improvement)
Run standard, same poor peak shape (earlier RT due to column trimming).
Visually inspect liner and injector seal for septum or other debris, none noted.
Install column2
Run standard, same poor peak shape as on column1. (valley between co-eluting peaks = 15%)
I'm not having any air leak issues, using 18/28/32 m/z ratios as diagnostic as well as argon leak testing the column fittings/union. Pressure is holding fine. Even with the poor peak shape, retention time is stable (except when 1 loop of column1 was removed of course).
The GC shares a He tank with a second GC and that instrument is running fine. Just trying to rule out a "bad" GC tank as well.
I don't have access to a new column, but will try a third used column that ran good too when last run. But it's like everything "fell off" after that good run, and I'm trying to figure out what else to look for.
Thanks,
Grant