The analyte comes out at RT 3.5 min in an isocratic run with 50mM phosphate buffer (pH2.5) 15% acetonitrile/water, C18 column.
I identify SNAC by full UV spectrum (from 190 to 350nm). Then perform quantitation. However, there is noise background around the RT of SNAC which affects qualitative identification as well as quantitation. I tried 2 approaches to try to separate the noise from analyte:
1. Different columns (C8, HILIC).
2. SPE cleanup.
1. I tried using different ACN concentrations with HILIC column but even nitrite, NAC, and SNAC failed to be separated in aqueous standards. Using C8 column does not improve the background noise separation.
2. I tried Bond elut, HILIC, MCX by conditioning the SPE cartridge, then load the urine sample. The flow-through was collected for analysis. However, polar interferences remain and background noise is still there.
I am now thinking of 2 approaches to separate the interference from analyte:
1. try hexyl phenyl column.
2. changing the ACN conc in mobile phase in C18 run.
Am I right to try out other column type and then mobile phase organic conc.?
Or is there any SPE I can use to clean up polar interferences from inorganic anions?
I am grateful for any comments and directions.
Thanks.