How much baseline noise should I be getting with an FID on an SRI 8610B GC? I've been getting some peaks greater than 5 mV then a steady rise to almost 90mv as it continues running just straight helium. I just took over the lab ad the equipment so I don't quite no the ins and outs yet.

Thanks
This is too much noise.

One of your gases may be contaminated.

Or someone put something into the injector they should not have.

Do you see a flame in your FID? You shouldn't.

best wishes,

Rod
fpcwsr,

A dirty jet might cause this although there is not much of a jet in the SRI. A bad electrical connector to the amplifier could also cause this, especially if the collector wire runs over some other electrical source. You might also try turning your ignitor voltage to off if this is one of the units that continuously runs the ignitor.

Best regards,

AICMM
3 posts Page 1 of 1

Who is online

In total there are 50 users online :: 2 registered, 0 hidden and 48 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Georguslat, Google [Bot] and 48 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry