what is the most common reasons for clogging of ESI probe

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

4 posts Page 1 of 1
i am running a LC-MS instrument. One day ESI capillary got clogged. I just wonder LC column should be very good filter for these particles. why LC column did not stop these particles?
anything that is soluble can get through the column; anything that is not volatile will dry out in the spray chamber and may block either the capillary linking the spray chamber to the vacuum part of the instrument, or occasionally the spray needle.

The usual suspects are salty buffers in the solvents, but repeated injections of concentrated samples can do the same thing.
Seconding what lmh said. It's probably a salt in your solvents. I have encountered this on my instrument.
But i do not salt in my buffer. only volatile buffer.
Seconding what lmh said. It's probably a salt in your solvents. I have encountered this on my instrument.
4 posts Page 1 of 1

Who is online

In total there are 26 users online :: 0 registered, 0 hidden and 26 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 26 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry