DR » Thu Aug 04, 2011 9:04 pm
1. Is there a way to tell how long the delay is? I know when I run single injections, there's about a 15s delay between what shows up on the detector and what shows up on the computer screen. Should I assume that is the actual delay or go conservative and give it a minute or two at the end of each injection? Just to reiterate, I allow for this delay by implementing a "next injection delay", correct?
I'm looking for a minute or two here.
2. I want to be certain that my instrument method is communicating to the autosampler correctly. Currently, my method looks like this (I have two buffers and am running a 0-100% gradient of buffer B over 30 minutes):
0minutes, 1ml/min, 100%A, 0%B, "curve" box blank
30minutes, 1ml/min, 0%A, 100%B, Curve 6
35minutes, 1ml/min, 0%A, 100%B, Curve 6
45minutes, 1ml/min, 100%A, 0%B, Curve 6
50minutes, 0.01ml/min, 100%A, 0%B, Curve 11
Does this method adequately indicate a "stop" signal that tells the autosampler to move to the next injection? If not, what am I missing?
This method is setup for gradient, re-equilibrate, go to low flow rate if no more injections are made. If your 45 minute line is to reset initial conditions, I'd change that time to 36 minutes, or even 35.5 minutes. That will get it back to starting conditions faster. So after making that change, what I'd do is tell Empower to collect data for about 40 minutes and add a next injection delay of 2 minutes. Your total run time would then be 42 minutes between injections and you would have 40 minutes of data for each injection. If this does not help, then we need to look at the cables you mention below.
3. In addition to the SAT/IN module, we have a pump module, which I presume does the same for the pumps that the SAT/IN does for the detector. There is a cable connected to the "relay" slot on the terminal strip on the back of the box. Currently, the other end of the cable is not connected to anything. Should it be and, if so, where?
There should be something going to the SAT/IN to tell it when an injection has been made. If there is a 2 lead wire coming from either the autosampler or controller from a spot that says "Inject/Start" (or something similar), it should run to the SAT/IN. This is what triggers the SAT/IN to start collecting data for the run time specified in your sample set method.