I used a C18 column to analyze a reaction mix consisting of N-acetylcysteine (NAC) and nitrite in HCl. S-nitroso-N-acetylcysteine is the derivatized product to be measured. But since I used an excess of derivatizing agent (NAC), the latter is present in about 1361-1461ppm in the mix.
1. Mobile phase used: 15% ACN in water with phosphate buffer.
2. isocratic
3. At 210nm, NAC peak appears at solvent front but the UV spectrum is non-specific.
No sign of contamination (of the non-specific UV spectrum) in C18 column even during wash with gradient of 5% ACN to 70%ACN.
The contamination problem occurred after I changed to HILIC column and ran some samples.
HILIC condition as follows:
1. MP A: 0.1% formic acid in water; MP B: 100% ACN
2.
Time (min) A% B% Flow (mL/min)
0 5 95 1
4 95 5 1
7 95 5 1
17 5 95 1
18 5 95 0
3. In blank sample after sample, I see a peak and a plateau of UV absorbance at 210nm.
I washed the whole system (including column) with different ACN concentrations, then 0.1% formic acid in IPA: water (1:1). Contamination still present, though halved in height after wash.
I disconnected the column. Then washed with 3M HNO3. Contaminant still there.
I tried no injection, the plateau is still there. I connected the purge valve directly to DAD, the plateau is still there.
I can’t figure out the problem.
Please help. Thanks in advance.