If I use 2 standards and inject them 3 times (total of 6 injections) upfront, calculate the response factor (e.g. area/wt) and then calc the overall RSD for the 6 std injections, doesn't this satify the USP system suitability criterion? Do I have to inject 6 injections of standard 1 & 2 (total of 12 injections) to meet USP criterion? I believe 2 standards injected 3 times is equivalent to the USP approach and therefore, I recommend USP to change the description in chapter <621>....."replicate injections of a standard preparation.....". Often, the system suitability standards are used for the sample calculation so the 2 standards injected 3 times approach is frequently used instead of a single standard injected 6 times approach.
Your response is appreciated.