Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
hii,
I've been working on my sample for its analysis on LC/MS. I am using a C18 column and when I run my sample, it seems to not bind the column. It looks like it is just flowing away than binding into the column. Could you please suggest me as to what I can do to solve this.

Thanks a lot.
Awaiting your reply.
First of all,try to know the nature of your molecule or analyte. C18 is non polar, may be your analyste is polar. Why dont you change the column from C18 to some polar column. There are plenty of polar columns to choose.Like HPLC, there is not mucg choice in LCMS to change the mobile phase. Just find out the nature of the analyte and then decide which column is best for it.
If you were a little more specific on what you were analyzing and how, we might be able to offer an advice. But without any info, we can only speculate.

Why not try HILIC?
3 posts Page 1 of 1

Who is online

In total there are 26 users online :: 1 registered, 0 hidden and 25 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 25 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry