lmh » Wed Aug 17, 2011 9:18 am
LC chemstation is totally different to GC chemstation, but smoothing is easy.
In the signal details box, if you are carrying out off-line data-processing, you will probably have some signals specified in your method. If not, when opening a file, tick the box marked "load using signal details". Now go to signal details, and transfer anything you want into the table of signals used in the method, which is towards the bottom of the signal details page. The available signals are in a drop-down box at the top.
Now move to the extreme right hand end of this table. You will find a column in the table for smoothing, where you can specify the algorithm, and another column where you can specify the number of points over which smoothing is carried out.
I STRONGLY recommend Savitsky-Golay smoothing rather than Gaussian. If you want to know why, try both... with S-G you can use a huge number of points to smooth over (11, 13...) without increasing peak-width.
There are a few hitches of which you should be aware:
(1) When you start a new calibration table, for some reason Chemstation will forget any smoothing information in the signal details box, so you'll have to restore this on opening your first calibration point.
(2) If you are using LC-MS and have SIM signals, they won't appear in the drop-down list to add to the method until you've extracted them using File - Extract ion chromatogram.